Code: Alles auswählen
Calibration of the Taiko™ Unit should be done whenever Software is updated,
if the Taiko™ Unit has been disassembled and/or if the CPU or Sensor Boards
have been replaced.
To calibrate a Taiko™ Unit, follow these steps:
1. Remove the Taiko™ Unit from its Faceplate mounting.
2. Set DIP Switches #1, #2 and #8 to ON.
3. Apply power to the Unit.
4. Confirm that the Front Panel LED is flashing WHITE.
5. Set DIP Switch #1 to OFF.
6. Confirm that the Front Panel LED is steady Lit GREEN.
7. Insert the KS-70 Reference Paper.
It will feed-in automatically and the GREEN LED will flash.
8. Set DIP Switch #8 to OFF.
The YELLOW LED will light and the Reference Paper will eject. the GREEN
LED will then flash.
9. Remove the Reference Paper from the Unit.
10.Set DIP Switch #8 to ON.
The YELLOW LED will light steady and then turn BLUE indicating calibration
is complete.
11. If the Front Panel LED turns RED, an error during calibration is indicated.
Isolate the problem by using the Error Code conditions listed in Table 7 on
page 18 of this Guide. Then re-calibrate the Taiko™ Unit by following this
procedure from Step 1 again.
Hab das jetzt einfach mal ausprobiert an einem Taiko mit nem normalen 50 Euroschein. Klappt auch das kalibrieren, aber klar, geht natürlich wohl nur mit dem Referenzschein.
Wäre nett, wenn mir jemand das Papier mal leihen/verkaufen könnte.Oder andere Quelle weiß.
Hätte gerne mal meine Taikos wieder in einem optimalen Zustand.